Mario, I assume that your prebake temperatures are missing a decimal point and they should be 50.0C and 90.0C. Assuming you can measure the light intensity at 405 nm the exposure should be about 50 mj/sq.cm per micron of resist, thus if you have 100 microns at an intensity of 23 mw/sq.cm (assuming 405 nm) the correct exposure would be (50 x100) / 23 equals about a 200 second exposure (I'm assuming here that the light intesity was measured at 405 nm). The pre bakes look fine, maybe a little long. The post exposure bake should be lower, near 65.0C, for about 90 to 180 seconds. The develop should be a two stage develop, first bath about 3/4 of the time required and second bath to rinse about 1/4 the total time required. I've used this for 20 to 100 u films with good results. Clif Clif >>> Mario Adamschik08/31 3:02 AM >>> Question for Discussion group: Subject: SU-8 50 Patterning I recently bought SU-8 50 resist and I tried to pattern the resist with the following parameter: Sample: Si wafer, about 2cm x 2cm Dehydration Bake 1h, 2300C Spinning: 700rpm 5s 2000rmp 35s 10min sheet relaxation pre bake 2min 500C Hotplate 30min 900C Hotplate Exposure: 9, 12, 15, 17, 19s (5 samples), 23mW/cm^2 post-bake 500C 2min 950C 5min slow cool down (2h) developing: 10min After developing the resolution was bad and there were a lot of rifts on the resist surface. Can anyone help me with optimized parameters to achieve sheet thicknesses of approx. 100um and 50...60um with SU-8 50 ?! Thank anyone very much in advance. Mario